2 scionix 38b57 comparison
Posted: 07 May 2017, 05:30
Hello everybody,
i want to share with you this comparison i made between 2 Scionix 38B57 E1 detectors,
the first has a bad resolution, around 8,5% @ 662KeV due to some wet in the crystal.
Recently I had the opportunity to refurbish another detetector, this time the things are different,
i obtained a good resolution of 7% both in Theremino MCA and BeqMoni.
The 2 probe was setted in the same way, with the same home made voltage divider of 120Mohm as shown below and both supplied with a 740 Volt bias.
In the third picture you can see the benefits of the new probe with improved resolution, especially on low-mid energies.