Scionix 38B57 calibration and linearity
Posted: 27 Sep 2019, 07:20
Hello everyone,
I'm really new to gamma spectroscopy and would like to share the results of my first few experiments. I welcome and appreciate any suggestions that you may have to improve my setup and operations. I wasn't sure if this was the right topic to post, please correct me if I'm wrong.
My detector is a refurbished Scionix 38B57 from iRad driven by a GS-USB-PRO. I house the detector in an improvised lead pit that I made with lead scraps. The shielded background count rate is 13.4 cps, a little over 4x reduction of the unshielded background.
I don't have a good calibration source yet, so I use a thoriated lantern mantle to do all my calibrations. I use PRA for all my measurements. At first I started acquiring spectra at 750 V and fairly short pulses, because that seemed the best way to enhance high energy peaks that were otherwise difficult to see. I soon discovered that this was not good at all, because it made the detector very non-linear above 900-1000 keV. Fortunately I found some posts on this forum that helped me understand what was going on, and I started experimenting with lower voltages and longer pulses. I read an advice from Steven to keep the K-40 1461 keV peak around 50 arbitrary units for best results, so I tried to do that.
I found that the best results in terms of linearity come with the pulse length at maximum setting and voltages of 650 V or below. Lower voltages make the detector more linear, but I had to raise the output volume to compensate. At 575 V I have a fairly good linearity up to 2600 keV, although the volume setting is at maximum and I had to add a 1.25x boost setting in PRA to keep 1461 keV around 50 arbitrary units.
Below are the results of my linearity experiments with various voltages. Each point in the plots is a major peak in the spectrum, the first plot has labels with the energies. The non-linearity always seems to appear beyond the 1588 keV Ac 228 peak.
Here is the thoriated mantle spectrum acquired at 575 V. For the interpretation I followed the spectra posted in this forum (thanks!) and elsewhere, along with some published energy lists. I'm not so sure about the 1588 keV Ac 228 peak, because I couldn't find a good published reference for that. Also, I'm not sure how to interpret the peaks at 2124 and 2303 keV. The first may be a single escape peak (2615 - 511 = 2104, well within the uncertainty), but I have no clue about the second. There might be a a couple of bumps at 209 and 277 keV from Ac 228 and Tl 208, but that seemed an over-interpretation to me.
I'm really new to gamma spectroscopy and would like to share the results of my first few experiments. I welcome and appreciate any suggestions that you may have to improve my setup and operations. I wasn't sure if this was the right topic to post, please correct me if I'm wrong.
My detector is a refurbished Scionix 38B57 from iRad driven by a GS-USB-PRO. I house the detector in an improvised lead pit that I made with lead scraps. The shielded background count rate is 13.4 cps, a little over 4x reduction of the unshielded background.
I don't have a good calibration source yet, so I use a thoriated lantern mantle to do all my calibrations. I use PRA for all my measurements. At first I started acquiring spectra at 750 V and fairly short pulses, because that seemed the best way to enhance high energy peaks that were otherwise difficult to see. I soon discovered that this was not good at all, because it made the detector very non-linear above 900-1000 keV. Fortunately I found some posts on this forum that helped me understand what was going on, and I started experimenting with lower voltages and longer pulses. I read an advice from Steven to keep the K-40 1461 keV peak around 50 arbitrary units for best results, so I tried to do that.
I found that the best results in terms of linearity come with the pulse length at maximum setting and voltages of 650 V or below. Lower voltages make the detector more linear, but I had to raise the output volume to compensate. At 575 V I have a fairly good linearity up to 2600 keV, although the volume setting is at maximum and I had to add a 1.25x boost setting in PRA to keep 1461 keV around 50 arbitrary units.
Below are the results of my linearity experiments with various voltages. Each point in the plots is a major peak in the spectrum, the first plot has labels with the energies. The non-linearity always seems to appear beyond the 1588 keV Ac 228 peak.
Here is the thoriated mantle spectrum acquired at 575 V. For the interpretation I followed the spectra posted in this forum (thanks!) and elsewhere, along with some published energy lists. I'm not so sure about the 1588 keV Ac 228 peak, because I couldn't find a good published reference for that. Also, I'm not sure how to interpret the peaks at 2124 and 2303 keV. The first may be a single escape peak (2615 - 511 = 2104, well within the uncertainty), but I have no clue about the second. There might be a a couple of bumps at 209 and 277 keV from Ac 228 and Tl 208, but that seemed an over-interpretation to me.