hi steven,
I just bought the gs-1100 A but I could not make decent spectra.
the spectra did not come because the tension was not correct.
if I put the gs at 9 to get 750v
the spectrum is ok.the voltage I measure it with a high-impedance multimeter, (1gohm) I measure the theremino and it's ok. What I asked was if there is a way to "tune" the tensions with the talking scale.

User avatar
gianlu383
Posts: 11
Joined: 15 Sep 2018, 18:35

Re: GS1100-a voltage out inaccurate

Post by gianlu383 » 16 Sep 2018, 01:14

thank you regards.

User avatar
Sesselmann
Posts: 1392
Joined: 27 Apr 2015, 11:40

Re: GS1100-a voltage out inaccurate

Post by Sesselmann » 16 Sep 2018, 09:39

Hi and welcome to the forum..

Sorry to hear that you are experiencing issues with the GS-1100A. To answer your question first, no there is no way to adjust the voltage on the GS-1100A model, the circuit uses a string of zener diodes to regulate the voltage, and the 12 way switch simply taps out the voltage between the zeners.

The circuit can only supply a small current, so it is essential to use a high impedance volt meter when testing, but I see you already understand this.

I test the GS-1100A on my work bench with a 1 Gohm, 100:1 divider and the voltages should be accurate to within 10 volts after a few minutes warm up.

It's very uncommon to hear of this problem, in fact I don't think anyone has brought this up before, but I guess it's possible for a diode to fail, or maybe I had a bad day and soldered one diode the wrong way :)

If you are confident there is a problem, I am happy to send you a new replacement board.

PS: We have a rule here on our forum that we use real names (at least first name), so please create a signature in the UCP, so members can address each other by real names.

Thanks.

Steven

User avatar
gianlu383
Posts: 11
Joined: 15 Sep 2018, 18:35

Re: GS1100-a voltage out inaccurate

Post by gianlu383 » 16 Sep 2018, 19:11

Thanks steven,
For more clarity I am attaching photos of my measurement system.
I did a test with the eberline esp2.
set at 700 vdc. I have a 60 vdc error that I know. measuring the gs1100
set at 7 I have 200 vdc less. add an instrumentation error of 20 vdc.
thank you. regards.
You do not have the required permissions to view the files attached to this post.

User avatar
Sesselmann
Posts: 1392
Joined: 27 Apr 2015, 11:40

Re: GS1100-a voltage out inaccurate

Post by Sesselmann » 16 Sep 2018, 20:54

Gianluka,

Thanks for images, it looks like we are out by 200V, which is consistent with a faulty or wrong soldered Zener diode. The first three Zeners are 200V each, which is why the GS-1100A srarts at 600V. I made up a new board today, will post it tomorrow.

BTW, as a mad fusioneer, I love your flux capacitor ;)

Steven

User avatar
gianlu383
Posts: 11
Joined: 15 Sep 2018, 18:35

Re: GS1100-a voltage out inaccurate

Post by gianlu383 » 17 Sep 2018, 00:25

ok, thanks steven,
congratulations for your work and for this beautiful forum.

PS: maybe one day you can make a flux capacitor ..... remember ... 1.21 gigowatt :-)))

User avatar
Sesselmann
Posts: 1392
Joined: 27 Apr 2015, 11:40

Re: GS1100-a voltage out inaccurate

Post by Sesselmann » 17 Sep 2018, 11:05

Gianluka,

Replacement board shipping today, check your PM for tracking number.

My best fusion energy output is around 5x10^6 neutrons/second which works out to around 3 microwatt ;)

Steven

User avatar
gianlu383
Posts: 11
Joined: 15 Sep 2018, 18:35

Re: GS1100-a voltage out inaccurate

Post by gianlu383 » 17 Sep 2018, 23:10

thank you so much steven.
I'll let you know as soon as it arrives.

User avatar
gianlu383
Posts: 11
Joined: 15 Sep 2018, 18:35

Re: GS1100-a voltage out inaccurate

Post by gianlu383 » 27 Sep 2018, 04:00

hello steven, today has arrived the new board.working very well!!!.I am very happy.I immediately made a bit of measurements.thanks to your disponability and assistance.
lu 176 calibration source

radio test

all ok. best regards
You do not have the required permissions to view the files attached to this post.

User avatar
Sesselmann
Posts: 1392
Joined: 27 Apr 2015, 11:40

Re: GS1100-a voltage out inaccurate

Post by Sesselmann » 27 Sep 2018, 11:55

Gianluka,

Great..., pleased to hear that it's working. Looks like the old board had a dead diode, or maybe one upside down diode (my fault :p ).

Great spectra, keep up the good work!