Fukushima Soil Sample "Reloaded" with both Cs134 Peaks
Posted: 26 May 2019, 07:23
Hi Folks,
Following our conversation (here https://www.gammaspectacular.com/phpBB3 ... a5fcda4c45) on Cs134 gamma peak at 604 keV detection in Fukushima soil sample I attempted a further experiment.
Steven suggested to use a Cs137 check source as a background in order to remove its gamma peak at 662 keV and get Cs134 first peak out of its slope.
That made me think about another possibility. I have no lead shield therefore every measurement I do involves a background subtraction, so when I test a sample I actually measure the sample + the background and then I subtract the latter.
So I thought “what if I subtracted the 662 keV peak together with my natural background”? In order to do so I needed to “build” an artificial composite background, adding to the natural background a Cs137 gamma peak of the same number of counts, relative to the natural peaks, of the sample + background spectrum. For instance, in the sample + background spectrum the counts ratio between Cs137 gamma peak centre and K40 gamma peak centre is approximately 15 to 1, so in my composite background I aimed to have the same ratio.
In order to have the right counts ratio between the 662 keV peak and the rest every 10 seconds of my composite background was made of roughly 9 seconds of just background and 1 second of Cs137 check source (9.25 kBq).
In the end I subtracted this artificial background from the sample+background spectrum and the 662 keV peak completely disappeared , leaving both Cs134 peaks well visible.
Both peaks are 5-6 keV lower than they should be, that might depend on a calibration drift (the whole test lasted about 24 hours). Anyway, I guess given the kind of measurement that’s acceptable.
The base of the 604 keV peak is slightly stretched towards lower energies, I wonder if it might be because of the presence of the other Cs134 peak at 569 keV (and in theory there's another one at 563).
The 604 keV ROI contains 28k counts, the 795 keV ROI contains 34k counts.
For obvious reasons they stand out more in the energy per bin view.
I have “something” in K40 zone. This is the third spectrum of this sample and something looking like K40 materialised the first and the third time around, but not the second (where the very weak peak was around 1400 keV, a bit too low even in case of a calibration drift).
Massimo
Following our conversation (here https://www.gammaspectacular.com/phpBB3 ... a5fcda4c45) on Cs134 gamma peak at 604 keV detection in Fukushima soil sample I attempted a further experiment.
Steven suggested to use a Cs137 check source as a background in order to remove its gamma peak at 662 keV and get Cs134 first peak out of its slope.
That made me think about another possibility. I have no lead shield therefore every measurement I do involves a background subtraction, so when I test a sample I actually measure the sample + the background and then I subtract the latter.
So I thought “what if I subtracted the 662 keV peak together with my natural background”? In order to do so I needed to “build” an artificial composite background, adding to the natural background a Cs137 gamma peak of the same number of counts, relative to the natural peaks, of the sample + background spectrum. For instance, in the sample + background spectrum the counts ratio between Cs137 gamma peak centre and K40 gamma peak centre is approximately 15 to 1, so in my composite background I aimed to have the same ratio.
In order to have the right counts ratio between the 662 keV peak and the rest every 10 seconds of my composite background was made of roughly 9 seconds of just background and 1 second of Cs137 check source (9.25 kBq).
In the end I subtracted this artificial background from the sample+background spectrum and the 662 keV peak completely disappeared , leaving both Cs134 peaks well visible.
Both peaks are 5-6 keV lower than they should be, that might depend on a calibration drift (the whole test lasted about 24 hours). Anyway, I guess given the kind of measurement that’s acceptable.
The base of the 604 keV peak is slightly stretched towards lower energies, I wonder if it might be because of the presence of the other Cs134 peak at 569 keV (and in theory there's another one at 563).
The 604 keV ROI contains 28k counts, the 795 keV ROI contains 34k counts.
For obvious reasons they stand out more in the energy per bin view.
I have “something” in K40 zone. This is the third spectrum of this sample and something looking like K40 materialised the first and the third time around, but not the second (where the very weak peak was around 1400 keV, a bit too low even in case of a calibration drift).
Massimo