Thorium mantle gamma spectrum with 2x2 inch Scionix NaI detector

Post by Nephrondoctor » 18 May 2015, 04:43

Thorium mantle gamma spectrum with my new (demo) 2x2” NaI detector from Scionix, GS1100 A and E-MU 0204
Attachments
Thorium mantle gamma spectrum with my new Scionix 2x2detector ...pdf
(1.13 MiB) Downloaded 831 times

rodorr
Posts: 12
Joined: 10 May 2015, 13:29
Contact:

Re: Thorium mantle gamma spectrum with 2x2 inch Scionix NaI detector

Post by rodorr » 29 May 2015, 03:42

What voltage was your GS-1100A set to when you obtained the Thorium spectrum?

Bob

rodorr
Posts: 12
Joined: 10 May 2015, 13:29
Contact:

Re: Thorium mantle gamma spectrum with 2x2 inch Scionix NaI detector

Post by rodorr » 02 Jun 2015, 00:15

I've answered my own question by paying attention to your document (800 Volts). Note to self, Read the included documents more closely,

Bob
WA7ZQR

User avatar
Sesselmann
Posts: 1374
Joined: 27 Apr 2015, 11:40
Location: Sydney
Contact:

Re: Thorium mantle gamma spectrum with 2x2 inch Scionix NaI detector

Post by Sesselmann » 02 Jun 2015, 07:44

Bob,

The PMT bias voltage varies from detector to detector, this is due to the difference in gain between PMT's. PMT's are essentially amplifiers with a gain in the order of a million, this gain is a function of the coating thickness on the cathode (the brown coating on the inside of the glass) and how efficiently it converts photons to electrons.

The PMT manufacturers test each PMT and supply a data sheet specifying the gain. When I order PMT's for spectrometry, I specifically ask for tubes with low gain, in the range 150 to 500 ma @ 650V, this works best with the GS.

In the last batch of PMT's I ordered, the gain varied from 337 ma to 474 ma.

Note also that the Pulse Height Resolution (PHR) also varies from one PMT to another, so if you have a bad detector it may not be all due to the crystal. There can be as much as a 2% variance in the PHR @ 662 keV in a brand new PMT. The manufacturers test each PMT in a dark test chamber with a standard NaI(Tl) crystal and record the PHR, which can be around 6.5% for an excellent PMT and 7.5% for a not so good one.

The resolution of a PMT is a function of how well the dynodes focus the electron beam as it multiplies itself down the set of dynodes. Badly focused electrons fail to make it all the way to the anode, and cause the final pulse to be slightly low.

Attached is a picture of a box and grid from a PMT which I broke accidentally.
ADIT PMT
ADIT PMT
photo.JPG (158.73 KiB) Viewed 6834 times
The dynodes are the parts that look lice little scoops, and I think that round disc on the side contains a getter, the getter is made from a porous active metal which sucks up any remaining gas molecules after the PMT has been evacuated. Each dynode is held apart by a series of ceramic rings.