Gamma spectrometry using eV CZT and PRA10

Share your spectra and discuss their features here
User avatar
Sesselmann
Posts: 1398
Joined: 27 Apr 2015, 11:40

Re: Gamma spectrometry using eV CZT and PRA10

Post by Sesselmann » 03 Feb 2016, 21:26

ba.png
Andrew,

Here is an old recording done with eV-CZT on PRA-3, the spectrum is Ba133 and the tall peak is at 81 Kev which is arb.u 0.61.

Line level at around + - 1.75V so divide this by 100 arb U and get 17.5 mA per arb.u,

As far as I recall the eV-CZT was connected directly into the line in port.

Steven

Note: PC line level and amplification can vary from sound card to sound card.
You do not have the required permissions to view the files attached to this post.

User avatar
Nephrondoctor
Posts: 18
Joined: 17 May 2015, 22:19

Re: Gamma spectrometry using eV CZT and PRA10

Post by Nephrondoctor » 05 Feb 2016, 00:37

Nephrondoctor wrote:
lodovico wrote:Alfred,
very Interesting the use of the sensor ev CZT. What about the low energy gammas ? Did you try with Americium ?
Lodovico
Yes, I tested Am241 too, but I am not sure which files I saved. I am now at work, will look into it this evening. Regards, Alfred

Enclosed two graphs showing 241Am, using CZT and using NaI(Tl). Resolution @59.5 keV appeared worse with CZT (30%) than with NaI (25%), in contrast to the results with Lutetium.
241Am PRA10 Scionix 2x2 inch 900V 0.2 0.5 0 1180K 1700 sec.JPG
241Am PRA10 CZT 150V EMU60% 0.2 30 0 1045K 1265 sec.JPG
You do not have the required permissions to view the files attached to this post.