Andrew,
Here is an old recording done with eV-CZT on PRA-3, the spectrum is Ba133 and the tall peak is at 81 Kev which is arb.u 0.61.
Line level at around + - 1.75V so divide this by 100 arb U and get 17.5 mA per arb.u,
As far as I recall the eV-CZT was connected directly into the line in port.
Steven
Note: PC line level and amplification can vary from sound card to sound card.
Gamma spectrometry using eV CZT and PRA10
-
Sesselmann
- Posts: 1398
- Joined: 27 Apr 2015, 11:40
Re: Gamma spectrometry using eV CZT and PRA10
You do not have the required permissions to view the files attached to this post.
-
Nephrondoctor
- Posts: 18
- Joined: 17 May 2015, 22:19
Re: Gamma spectrometry using eV CZT and PRA10
Nephrondoctor wrote:Yes, I tested Am241 too, but I am not sure which files I saved. I am now at work, will look into it this evening. Regards, Alfredlodovico wrote:Alfred,
very Interesting the use of the sensor ev CZT. What about the low energy gammas ? Did you try with Americium ?
Lodovico
Enclosed two graphs showing 241Am, using CZT and using NaI(Tl). Resolution @59.5 keV appeared worse with CZT (30%) than with NaI (25%), in contrast to the results with Lutetium.
You do not have the required permissions to view the files attached to this post.