I am just getting started experimenting with a modified Bicron 1.5M2.25/1.5L NaI(Tl) probe, using Theremino hardware and software. Just getting started with gamma spectroscopy other than using a Radiacode.
I was told that perhaps I could get a more linear response, not needing as much adjustment with the ThereminoMCA linearizer, if I lowered the voltage to the probe. I lowered from about 730 volts to 650 volts and it appeared that I lost some peaks when testing Am241 from a smoke detector. I did find that the linearizer did not need to be adjusted as far from center positions with the lower voltage.
Is such a thing normal or am I doing something wrong?
Any suggestion on what voltage should be used?
NaI(Tl) Gamma Scintillation detector Voltage
-
GetSirius
- Posts: 4
- Joined: 06 Dec 2024, 06:19
NaI(Tl) Gamma Scintillation detector Voltage
You do not have the required permissions to view the files attached to this post.
-
ColoRad-o
- Posts: 64
- Joined: 18 Oct 2019, 08:27
Re: NaI(Tl) Gamma Scintillation detector Voltage
I ultimately gave up on seeking for very linear response (I assume you mean: how the location of gamma peaks change as you change bias?
Here are several plots that may answer some of your questions. Perhaps the file Plateaux.pdf is the most important--ignore the Ludlum data (it was a bad detector). You could argue that you should select the bias voltage that keeps the response of the detector system (here, in counts per second, if I remember correctly, for some damn peak or another, presumably using a gamma source) as insensitive to bias voltage as possible. The right hand side of the plateaux plot is the derivative of the response with respect to bias.
There was a gain boost within PRA (which I used for all of the above) that has a large impact on locations in the (arbitrary unit) spectrum .
Hope this helps!
DMW
Here are several plots that may answer some of your questions. Perhaps the file Plateaux.pdf is the most important--ignore the Ludlum data (it was a bad detector). You could argue that you should select the bias voltage that keeps the response of the detector system (here, in counts per second, if I remember correctly, for some damn peak or another, presumably using a gamma source) as insensitive to bias voltage as possible. The right hand side of the plateaux plot is the derivative of the response with respect to bias.
There was a gain boost within PRA (which I used for all of the above) that has a large impact on locations in the (arbitrary unit) spectrum .
Hope this helps!
DMW
You do not have the required permissions to view the files attached to this post.
-
GetSirius
- Posts: 4
- Joined: 06 Dec 2024, 06:19
Re: NaI(Tl) Gamma Scintillation detector Voltage
Thank you for the information. It is helpful. I am finding that my probe is very sensitive to bias voltage. I am trying to get back to the arrangement that produced the "730V" graph with the nice low energy peak, finding that just 10-20 volts at ~730V makes a noticeable difference in peak location, throwing calibration off.
I am not sure the Theremino hardware I am using is all that (temperature?) stable, or my meter system is not stable. In 24 hours with probably a 10-15F drop in room temperature, I measure a 35 volt increase in PMT bias.
I understood soon after first starting to experiment with this probe that peaks move, gain varies, when I change PMT voltage. My main concern when first making this post was how my low energy peak disappeared when I lowered the bias from around 730 to 650 volts. So far, I have not been able to get back to that nice sharp 26KeV peak, it comes up rather wide while the higher peak stays sharp.
I think one thing I need to do is replace the voltage adjust pot on the Theremino board with a 10 turn pot to make adjustment easier and maybe more stable, it is so fiddly with the one turn pot.
I am not sure the Theremino hardware I am using is all that (temperature?) stable, or my meter system is not stable. In 24 hours with probably a 10-15F drop in room temperature, I measure a 35 volt increase in PMT bias.
I understood soon after first starting to experiment with this probe that peaks move, gain varies, when I change PMT voltage. My main concern when first making this post was how my low energy peak disappeared when I lowered the bias from around 730 to 650 volts. So far, I have not been able to get back to that nice sharp 26KeV peak, it comes up rather wide while the higher peak stays sharp.
I think one thing I need to do is replace the voltage adjust pot on the Theremino board with a 10 turn pot to make adjustment easier and maybe more stable, it is so fiddly with the one turn pot.
-
Sesselmann
- Posts: 1392
- Joined: 27 Apr 2015, 11:40
Re: NaI(Tl) Gamma Scintillation detector Voltage
Chuck,
Right, linearity generally improves with lower voltage, not sure if this is caused by the PMT or the opamp, but you can test with a Thorium source and find a good compromise.
Here is a spreadsheet I made.
Steven
Right, linearity generally improves with lower voltage, not sure if this is caused by the PMT or the opamp, but you can test with a Thorium source and find a good compromise.
Here is a spreadsheet I made.
Steven
You do not have the required permissions to view the files attached to this post.