I've had my share of issues and questions with detectors, as detailed in my other threads. I've been trying to find "the one" detector that will be the "flagship" for my gamma spectrometry work and the one my "final" shield will be optimized around.
I thought I had found it, a new old-stock 3x3" Bicron at a price I could afford. It arrived shiny, well packed, and had a presence that meant business. I had it on my desk for several weeks as I first selected, and then waited for a voltage divider socket. At last the two met, and I excitedly watched the spectrum develop, and it was...
Garbage. At least compared to my other detectors, one of which (as detailed in another thread) is pretty beat up. Compared to my expectations, it takes more voltage and gain, has a lower count rate, and has very (very) wide peaks.
Since I had a new detector and a new divider base, I wasn't sure where the problem might be. I was able to borrow a different PMT base and that gives similar results, so I'm thinking it's the detector.
Here's a comparison of the same sample (a relatively hot piece of Trinitite) done with 3 detectors. The tests aren't consistent in shielding or length, but they are pretty typical for each detector:
Red: The brand new 3x3" detector
Blue: Old and degraded 5x4" well detector (The crystal is cloudy and has at least one crack and the light pipe is partially delaminated from the crystal).
Orange: A well used Ludlum 44-10 2x2" detector (old and well used enough that that you can actually feel something shift around inside as you move it side to side).
Is there anything else I should check?
- I feel like I've eliminated cables by swapping them between detectors.
- I'm using Theremino MCA and the pulses look OK to me, though there's some noise (probably due to the increased gain).
- I feel like the voltage divider is eliminated due to the substitution.
- I've tried higher and lower voltages (+/-100V from 1050V, which seems optimal)
Is there some magical setting for the Theremino MCA baseline test parameters that will improve resolution?
Do detectors like this typically have magnetic shielding built in?
Do I need to ground the case somehow?
Have you seen abnormally poor resolution before? Did you figure out the cause?
Are my expectations too high? Is this typical for a 3x3 detector?
Thanks in advance for any help or advice.
Mike S.