Hello!
Please let me know what will I see on GS-1100-PRO voltage meter - output voltage from the voltage converter or actual voltage on my detector? In the unit there should be a load resistor which will cause difference in voltage between output of HV converter and output connector.
Thank you!
GS-1100-PRO - What will I see on the voltage meter?
-
trigatronix
- Posts: 4
- Joined: 07 Jan 2017, 09:33
- Contact:
- Sesselmann
- Posts: 1374
- Joined: 27 Apr 2015, 11:40
- Location: Sydney
- Contact:
Re: GS-1100-PRO - What will I see on the voltage meter?
Alexander,
Yes there is a load resistor in the GS-1100-Pro, so you need to consider this as part of your voltage divider, meaning the actual voltage on the PMT may be somewhat less than what it says on the display.
When I calibrate the display I use a 1 G Ohm divider, so if your divider os only 10M Ohm then the voltage on the pmt might be V = n(10/11)V
PS: Please change your username or add a signature in the footer, so we know who you are when you post, I had to look up your name in the introductions forum.
Thanks..
Steven
Yes there is a load resistor in the GS-1100-Pro, so you need to consider this as part of your voltage divider, meaning the actual voltage on the PMT may be somewhat less than what it says on the display.
When I calibrate the display I use a 1 G Ohm divider, so if your divider os only 10M Ohm then the voltage on the pmt might be V = n(10/11)V
PS: Please change your username or add a signature in the footer, so we know who you are when you post, I had to look up your name in the introductions forum.
Thanks..
Steven
Steven Sesselmann | Sydney | Australia | https://gammaspectacular.com | https://beejewel.com.au | https://www.researchgate.net/profile/Steven-Sesselmann
Who is online